    Question Blood glucose monitoring - which device do you use?

    HERE - READ WHY THIS ADVERT IS DISPLAYED!
    We currently use the Freestyle glucose monitoring device (a variant of this one http://www.abbottdiabetescare.com/fr...ng-system.html) but feel frustrated about it.

    This monitoring system was evaluated at Karolinska about ten years ago (http://www.ncbi.nlm.nih.gov/pubmed/16299875) and the correlation with our previous "gold standard" - venous blood sugars measured with the Hemocue machine - was found to be good.

    Now, we often see that low values obtained with the Freestyle are commonly normal when venous samples are measured by the regular lab or with Hemocue.

    What equipment do you use for blood glucose monitoring in your unit?

    Does anyone have experience from micro-dialysis of newborns/preterms?
    Stefan Johansson, MD PhD
    Karolinska university hospital
    Stockholm, Sweden

  2. #2
    Hello Stefan!

    I do not have experience with micro-dialysis but I am studying (pilot study) a subcutaneous glucosemonitor (Medtronic) and having good results with this device in late preterm and newborn babies. (Correlation between HemoCue and the Medtronic device (CGMS))

    Greetings from Huddinge

    Dirk
    Dirk Wackernagel, MD, Dr. med.
    Neonatal care K77-K79
    Astrid Lindgren Children's hospital
    Stockholm Huddinge
    Sweden

  3. #3
    99nicu Team Stefan Johansson's Avatar
    Join Date
    Apr 2006
    Location
    Stockholm, Sweden
    Posts
    1,062
    Blog Entries
    60
    Quote Originally Posted by wackdi View Post
    ...(pilot study) a subcutaneous glucosemonitor (Medtronic)...
    Sounds promising. Keep us (in Danderyd and in the world) posted!
    Stefan Johansson, MD PhD
    Karolinska university hospital
    Stockholm, Sweden

  4. #4
    99nicu Team Stefan Johansson's Avatar
    Join Date
    Apr 2006
    Location
    Stockholm, Sweden
    Posts
    1,062
    Blog Entries
    60
    I just found out that we do not longer have the device that was tested here in the early 2000s (http://www.ncbi.nlm.nih.gov/pubmed/16299875), but a "updated" version...

    So, the truth is that we are using a non-validated device. Which does not seem to work for neonates.

    We'll speak to Abbott (the manufacturer) but we are now really looking into the market for a new method for blood glucose monitoring.
    Stefan Johansson, MD PhD
    Karolinska university hospital
    Stockholm, Sweden

  5. #5
    HERE - READ WHY THIS ADVERT IS DISPLAYED!
    Hi

    I felt that glucometers that we use in NICU basically work in the same way , regardless of the make . However , I may be wrong
    I am presently in retrieval services , and hence use the iSTAT machine for getting the entire electrolytes , blood gas and glucose . This is fantastic as far as retrieval services are concerned . Even in NICU , where we do regular blood gases , the blood glucose value that appears on the gas result is often relied upon . It seems its reliable to an extend .. And the amount of blood required for glucometer and for ISTAT , capillary blood gas dont differ much ....

    Regards
    Gopan

Similar Threads

  1. Glucose before enteral feeding?
    By shiwenxia in forum nutrition
    Replies: 8
    Last Post: 13th June 2007, 14:07
  2. Frequency of blood gas monitoring
    By doc4tots in forum pulmonary disorders and ventilatory support
    Replies: 1
    Last Post: 17th December 2006, 13:10

Tags for this Thread

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •